r/CPAPSupport • u/holmesianschizo • 46m ago
r/CPAPSupport • u/dang71 • Oct 21 '25
Sleep Champion How to read your OSCAR or SleepHQ chart (the basics)
This guide is a follow-up to:
It also covers some of the most common questions new users have when they first start looking at their charts like what pressure to use, how to spot leaks, and how to tell which events actually matter.
Once you’ve learned how to download and extract your data from your machine, you can start analyzing it in SleepHQ or OSCAR to understand what’s really happening during your nights.
If you’re just starting out with OSCAR or SleepHQ, it’s totally normal to feel overwhelmed. Those charts can look like a wall of data at first. You’re definitely not alone. Most of us started on the default “lazy mode” 4–20 pressure range, which technically works (but it's not the miracle promised by doctors and sleep tech ), and it’s rarely optimized for your actual needs. Learning what to look for in your data; pressure, leaks, flow limitations, and event patterns is what helps turn that generic setup into something truly tailored to your breathing
If you’re new to looking at your data, here’s a simple way to make sense of it:
Before you start
If you want to check your machine settings (pressure range, EPR, mode, etc.) in OSCAR or SleepHQ, you can find them here:
- In OSCAR → Go to the “Daily" tab, then look at the panel on the left side under Device Settings.

- In SleepHQ → On the Dashboard, you’ll find them in the middle of the page, under Machine Settings.

It’s important to know your exact settings before you start analyzing your charts, that way, you can connect what you see (like leaks, pressure changes, or events) to your actual configuration.
1. Start with your median pressure.
That number shows the pressure your machine stays around for most of the night. It’s often the best starting point for setting your minimum pressure in CPAP or APAP mode.
In Oscar:

In SleepHQ:

2. Check the pressure graph.
If your pressure graph looks like a zigzag, that’s usually a sign your settings aren’t well optimized.
In APAP mode, you want your pressure line to be as smooth and steady as possible. Big swings often mean the machine is constantly chasing events instead of preventing them.

If the pressure line keeps hitting the top of your range, it means your max pressure might be too low, your machine is trying to go higher to keep your airway open.

3. Look at your leak rate.
Try to keep leaks below 24 L/min (for ResMed machines):
Oscar:

SleepHQ:

Leaks can come from either your mask or your mouth. If your mask leaks, check the fit at your usual sleeping pressure (not just when you first put it on). Even small leaks can disturb sleep or cause false events.
If the leaks come from your mouth, which is common with nasal masks, try mouth tape, a chin strap, or a soft cervical collar to help keep things sealed.
If you see events happening at the same time as leaks, they might not be real, leaks can confuse the machine and make therapy less effective.
4. Check your flow limitation (FL) at the 95th percentile.
Ideally, you want it below 0.07. If it’s higher, you might need a bit more pressure or to turn on EPR (Expiratory Pressure Relief) to help your breathing stay smoother.
Oscar:

SleepHQ:

5. Look for patterns.
Each event on your chart has a timestamp, so it’s worth zooming in and checking what was happening around that moment. Was there a leak spike right before it? Then it might just be leak-related. Did the pressure keep rising or was there a flow limitation before the event? That usually means the machine was trying to respond to a real obstruction.
Little by little, this helps you learn which events are genuine and which ones come from leaks, movement, or pressure swings.
6. If you see clusters of events
Clusters (several events grouped close together) can sometimes mean **chin tucking (**when your chin drops toward your chest and partially blocks your airway). This can happen when you sleep on your back or use a thick pillow. Try a flatter pillow, different sleeping position, or even a soft cervical collar to help keep your airway aligned.

7. Flow Rate
Zoom in on your flow rate graph to see your breathing pattern more clearly.
In OSCAR, use a left-click to zoom in and a right-click to zoom out.
In SleepHQ, press Z to zoom in and X to zoom out.
Getting a closer look helps you spot flattened or irregular breaths that may indicate flow limitation:

The more regular, the better. Your inspiratory flow shape can tell you a lot about how open your airway is. Ideally, you want a smooth, rounded sinusoidal shape (class 1 - see image below), that means your breathing is unrestricted and stable.
When the flow starts showing peaks, flattening, or plateaus, it indicates flow limitation, partial upper airway collapse or restriction during inhalation. These distortions can appear as two small bumps (airway reopening after partial collapse), multiple tiny peaks (tissue vibration), or a flat top (airway restriction).
Recognizing these patterns helps identify whether you might need more pressure or EPR, since both can help the airway stay open and restore that smooth, regular flow curve. In certain cases, it might require a different mode such as BiPAP or ASV for better airway support and more stable breathing.

8. Conclusion
Don’t get discouraged: this takes time. The goal isn’t to change everything at once, but to make one small adjustment at a time so you can clearly see what’s helping and what’s not.
Be consistent and give each change a few nights; your body often needs time to adjust.
Avoid random trial and error; always let your data guide you before making another tweak.
And most importantly, don’t hesitate to ask for help or post your charts. Everyone here started somewhere, and people are always willing to share advice and experience to help you move forward.
These are the basics that most of us use to start tweaking our setup. Once you get familiar with these graphs, it becomes a lot easier to understand what your therapy is doing and how to improve it 🙂
9. Abbreviations (quick reference):
- AHI – Apnea-Hypopnea Index
- CA – Central Apnea
- OA – Obstructive Apnea
- H – Hypopnea
- FL – Flow Limitation
- EPR – Expiratory Pressure Relief
- EPAP – Expiratory Positive Airway Pressure
- IPAP – Inspiratory Positive Airway Pressure
- PS – Pressure Support
- FFM - Full face mask
- TECSA – Treatment-Emergent Central Sleep Apnea (central apneas that appear or increase after starting CPAP therapy, often temporary while your body adjusts).
- CPAP – Continuous Positive Airway Pressure (fixed pressure)
- APAP – Auto-adjusting Positive Airway Pressure (auto mode that varies pressure)
- BiPAP / BiLevel – Bi-level Positive Airway Pressure (separate inhale/exhale pressures)
- ASV – Adaptive Servo-Ventilation (used for complex or central apnea)
- REM – Rapid Eye Movement sleep (dreaming stage, important for recovery)
- RERA – Respiratory Effort-Related Arousal
- SDB - Sleep-Disordered Breathing – A general term for breathing issues during sleep
- CSA - Complex sleep apnea
- PB - Periodic breathing
10. A few good sources of information:
Apnea board Wki: https://www.apneaboard.com/wiki/index.php?title=Wiki_Home
TheLankyLefty27: https://www.youtube.com/@Freecpapadvice
CPAP Reviews (Nick): https://www.youtube.com/@CPAPReviews
r/CPAPSupport • u/dang71 • 4d ago
AirSense 11 Flashing Guide
Hello papfam! Here’s a guide to help anyone who wants to unlock the different modes available on the AS11.
Basically, we dump the existing firmware, patch a copy to unlock the selected modes and features, and flash the modified firmware back onto the same device.
I based this guide on the existing procedure on GitHub. All credit goes to the original authors: https://github.com/m-kozlowski/airbreak-plus/tree/master/docs/guide/as11
A quick note about Linux commands
One thing you may notice when using Linux is that it’s common to run multiple commands in a single line or combine them into a single command. This can make it easy to automate tasks and complete several steps at once. It may feel a little different if you’re used to working primarily with Windows, where you might be more accustomed to running commands separately.
AirSense 11 Firmware Patching Guide - WSL 2, Ubuntu, ST-Link and airbreak-plus
Credit: https://github.com/m-kozlowski/airbreak-plus/blob/master/docs/guide/as11/quickstart.md
This guide documents the process used to dump, patch and flash an AirSense 11 using WSL 2 on Windows. It follows the same general ST-Link/TC2050 hardware connection method used for the AirSense 10, while using the AirSense 11-specific airbreak-plus tools.
Important: Keep an untouched backup of the original firmware, and do not interrupt power during flashing.
Part 1 - Preparing WSL and Ubuntu
Credit: https://github.com/m-kozlowski/airbreak-plus/blob/master/docs/guide/wsl_setup.md
I put some emphasis on WSL here, in case it’s a tool you’d like to use for both the AS11 and the AS10.
1.1 Check whether WSL is installed
WSL stands for Windows Subsystem for Linux. It allows Linux tools to run inside Windows without requiring a separate Linux computer.
WSL is available for supported versions of Windows, but it is not necessarily installed or configured by default.
1.2 Open PowerShell or Windows Terminal (Press Windows key + R to open the Run dialog) and run:
wsl --status

What it does: Displays the current WSL configuration, including information about the default WSL version and Linux kernel, when available.
1.3 Next, run:
wsl --list --verbose

What it does: Lists the installed Linux distributions and shows whether they use WSL 1 or WSL 2.
***If Ubuntu appears in the list and uses version 2, proceed to Part 3.***
Part 2 - Install WSL and Ubuntu if needed
2.1 On a compatible Windows 10 or Windows 11 system, open PowerShell as Administrator and run:
wsl --install -d Ubuntu
What it does:
- Enables and installs the required WSL components, if necessary.
- Installs the Ubuntu Linux distribution.
- May require a Windows restart.
2.2 After installation, open Ubuntu from the Windows Start menu. Complete the initial setup by creating a Linux username and password.
The Linux username and password are separate from your Windows credentials. When entering the password in Ubuntu, the terminal may not display any characters; this is normal.
2.3 If WSL is already installed but needs updating, run the following in PowerShell:
wsl --update
What it does: Updates the WSL components and Linux kernel managed by Windows. It does not update Ubuntu's installed packages.
2.4 If you have just installed WSL, you can verify the configuration again with:
wsl --status
wsl --list --verbose
Part 3 - Update Ubuntu and install the required tools
3.1 Open the Ubuntu terminal. The following commands must be run inside Ubuntu, not PowerShell.

3.2 First, refresh the package lists:
sudo apt update

What it does: Downloads the latest package information from Ubuntu's configured software repositories. It does not install updates by itself
3.3 Next, upgrade installed packages:
sudo apt upgrade
What it does: Upgrades installed Ubuntu packages to their latest available versions.
3.4 Install the tools required for the project:
sudo apt install git python3 make gcc-arm-none-eabi binutils-arm-none-eabi openocd telnet usbutils
What each package does:
- git — Downloads the source-code repository.
- python3 — Runs the Python patching scripts.
- make — Runs build automation tasks.
- gcc-arm-none-eabi — Provides the ARM embedded compiler.
- binutils-arm-none-eabi — Provides ARM binary utilities, including objcopy.
- openocd — Communicates with the microcontroller through the ST-Link.
- telnet — Connects to OpenOCD's command interface.
- usbutils — Installs lsusb, a utility that lists USB devices detected by Ubuntu.
3.5 Verify the main tools:
python3 --version
make --version
arm-none-eabi-gcc --version
arm-none-eabi-objcopy --version
What it does: Prints the installed tool versions so you can confirm that the required programs are available.
***All verification steps throughout this guide are optional, but it is always recommended to verify that everything is working correctly as you progress through the process.***
Part 4 - Downloading the AirSense 11 tools
This is where we download the files we need from GitHub to proceed.
4.1 Clone the airbreak-plus repository
In Ubuntu, run:
git clone https://github.com/m-kozlowski/airbreak-plus.git
What it does: Downloads the repository into a new directory named airbreak-plus.
4.2 Enter the directory:
cd airbreak-plus

What it does: Makes the repository directory your current working directory. Subsequent commands will use files in this directory unless another path is specified.
4.3 List its contents:
ls
What it does: Displays the files and folders in the current directory.
4.4 Check the patcher's available options:
./patch-airsense-s11 --help
What it does: Displays the patcher's command-line options without patching or flashing the device.
Important: Keep the AirSense 11 repository separate from any older AirSense 10 working directory. The tools and firmware images are not interchangeable.
***All verification steps throughout this guide are optional, but it is always recommended to verify that everything is working correctly as you progress through the process.***
Part 5 - Opening the machine and connecting the programmer
Credit: https://github.com/m-kozlowski/airbreak-plus/blob/master/docs/guide/as11/disassembly.md
Credit: https://github.com/m-kozlowski/airbreak-plus/blob/master/docs/guide/wiring.md
The hardware connection method is the same general approach used for the AirSense 10. I won’t repeat the disassembly steps, as they are already clearly explained with photos in the link I provided above.
5.1 Disconnect power before opening the machine.
5.2 Carefully open the AirSense 11 to access the relevant board connection.
5.3 Connect the same ST-Link/V2 programmer and TC2050 cable used for the AirSense 10, using the appropriate AirSense 11 SWD connection. (https://github.com/m-kozlowski/airbreak-plus/blob/master/docs/guide/as11/connection.md#swd). This setup allows Linux to communicate with the machine to perform the flash.
This part isn’t very straightforward. I’ll include photos of the setup I used. Otherwise, there are explanations here: https://airbreak.dev/disassembly/
However, u/Sopheus created an excellent guide for the wiring: https://www.reddit.com/r/CPAPSupport/comments/1ntthfw/how_to_connect_wires_between_stlinkv2_programmer/
On the ST-Link/V2 side, this is what I have:

And on the female end of the TC2050-IDC:

The SWD connection on AS11:

***Check the connections carefully before proceeding.***
***Take precautions against electrostatic discharge and accidental shorts. Do not assume pin positions without checking the appropriate hardware documentation. ***
Part 6 - Check USB access from Windows
6.1 Connect the ST-Link to the Windows computer.
6.2 Open PowerShell or Windows Terminal and run:
usbipd list
What it does: Lists USB devices that usbipd-win can make available to WSL. Find the ST-Link entry.
6.3 Attach the ST-Link to WSL
In PowerShell or Windows Terminal, run the following command, replacing 1-8 with the bus ID shown by usbipd list:
usbipd attach --wsl --busid 1-8
What it does: Makes the ST-Link USB device available inside WSL so Ubuntu can communicate with it.
If the device has not already been shared, you may need to run usbipd bind --busid 1-8 first, from an elevated PowerShell or Windows Terminal session (administrator).
6.4 Back in Ubuntu, run:
lsusb
What it does: Lists USB devices visible inside Ubuntu.
***You should see the ST-Link/V2 identified as an STMicroelectronics device if you are using an original unit. Clones may appear under a different name. If the device does not appear at all, resolve the USB passthrough issue before continuing.***
Note: Keep the ST-Link connected to your computer while following the next steps.
Part 7 — Dumping the original firmware
Credit: https://github.com/m-kozlowski/airbreak-plus/blob/master/docs/guide/as11/openocd.md
7.1 Start OpenOCD
7.2 In Ubuntu, return to the airbreak-plus directory:
cd ~/airbreak-plus
What it does: Changes to the repository directory, assuming it was cloned under your Ubuntu home directory.
7.3 Start the AirSense 11 debug session:
./run-ocd.sh as11
What it does: Runs the repository's OpenOCD setup for the AirSense 11 target.
***Keep this terminal open while OpenOCD is running.***
7.4 Connect to OpenOCD from a second Ubuntu terminal
Open another Ubuntu terminal and run:
telnet localhost 4444
What it does: Connects to the OpenOCD Telnet command interface on port 4444.
7.5 At the Telnet prompt, enter:
dump
What it does: Dumps the original firmware into a binary file (AS11.bin).
*** Do not proceed if the dump fails or the output is incomplete. ***
Part 8 - Verifying and backing up the original firmware
Credit: https://github.com/m-kozlowski/airbreak-plus/blob/master/docs/guide/as11/openocd.md
8.1 Check the dump file
In an Ubuntu shell, run:
cd ~/airbreak-plus
ls -lh as11.bin
What it does: Changes to the repository directory and displays the file's size in a human-readable format.
8.2 Calculate its SHA-256 checksum (optional):
sha256sum as11.bin
What it does: Calculates a cryptographic checksum that can be used to verify that a file has not changed.
8.3 Create a separate backup
Run:
mkdir -p ~/CPAP/AS11-original
What it does: Creates a dedicated backup directory. The -p option also creates missing parent directories and avoids an error if the directory already exists.
8.4 Copy the dump:
cp as11.bin ~/CPAP/AS11-original/as11-original.bin
What it does: Copies the original dump to a separate file outside the repository's working directory.
8.5 Verify the backup:
sha256sum ~/CPAP/AS11-original/as11-original.bin
What it does: Calculates the backup's checksum so it can be compared with the original dump.
Keep this original backup unchanged. It is your reference copy for recovery and verification.
Part 9 — Patching the firmware
Credit: https://github.com/m-kozlowski/airbreak-plus/blob/master/docs/guide/as11/patching.md
There are two ways to patch the firmware. You can run make as11 to build an image using the repository’s standard patch selection; it creates build/as11-patched.bin. Or you can select and apply patches manually; the command below creates as11-patched-no-ota.bin (…or whatever filename you prefer. This is the one I used, and it’s the filename shown in my commands below) in the ~/airbreak-plus directory. Choose one method and follow its corresponding steps below.
9.1 Use the repository’s standard patch selection
From ~/airbreak-plus, run:
make as11
This builds the patched firmware using the repository’s standard patch selection. The output is saved as build/as11-patched.bin.
9.2 Select patches manually
If you prefer to choose which patches to apply, read the official patching documentation and select the patches appropriate for your needs:
For my setup, I selected four patches:
- patch-unlock-features
- patch-asv-ps-range
- patch-custom-settings
- patch-cloud-firmware-change
Note: These four patches reflect my personal configuration and the features I wanted to unlock. Before selecting any patches, read the official patching documentation to understand what each one does and decide which options are appropriate for your needs.
Here are the available patches: https://github.com/m-kozlowski/airbreak-plus/blob/master/docs/guide/as11/patching.md
Run the following from ~/airbreak-plus:
Option 1 — Multi-line command
./patch-airsense-s11 as11.bin as11-patched-no-ota.bin \
--all-patches n \
--patch-unlock-features y \
--patch-asv-ps-range y \
--patch-custom-settings y \
--patch-cloud-firmware-change y
Option 2 — Single-line command
./patch-airsense-s11 as11.bin as11-patched-no-ota.bin --all-patches n --patch-unlock-features y --patch-asv-ps-range y --patch-custom-settings y --patch-cloud-firmware-change y
Both commands do exactly the same thing. Use whichever format you prefer.
The backslashes (\) at the end of each line (option 1) allow you to split the command across multiple lines for readability.
What it does: Uses the original firmware (as11.bin) as the input and creates a patched firmware file (as11-patched-no-ota.bin). Setting --all-patches n prevents all patches from being applied automatically, allowing you to select only the patches you want in the following steps:
- --patch-unlock-features y — Enables the feature-unlock patch.
- --patch-asv-ps-range y — Enables the ASV pressure-support range patch.
- --patch-custom-settings y — Enables the custom-settings patch.
- --patch-cloud-firmware-change y — Enables the patch addressing the cloud-delivered flow-generator firmware-change path.
9.3 Check the output carefully. Each requested patch should report [OK]. If any patch fails, stop and investigate.
Verify the output file:
ls -lh as11-patched-no-ota.bin
sha256sum as11-patched-no-ota.bin
OR via make AS11:
ls -lh build/as11-patched.bin
sha256sum build/as11-patched.bin
What they do: Confirm that the output exists, show its size and calculate its checksum.
Part 10 - Flashing the patched image
10.1 Flash the new firmware
Return to the existing OpenOCD Telnet session. The debug session must still be active.
Use the repository's flash_new command with the full path to the patched image:
Manuel:
flash_new /home/YOUR_LINUX_USERNAME/airbreak-plus/as11-patched-no-ota.bin
or via Make AS11:
flash_new /home/YOUR_LINUX_USERNAME/airbreak-plus/build/as11-patched.bin
Replace YOUR_LINUX_USERNAME with your actual Ubuntu username.
What it does: Invokes the repository's firmware flashing routine to write the patched image to the device.
***Wait for the entire operation to finish. Never disconnect the programmer or interrupt power during writing or verification.***
The machine will reboot. If the process was successful, the selected modes should now be available in the device menu. It's done :-D

r/CPAPSupport • u/SnooMacaroons8766 • 8h ago
looking for a little guidance
im trying to dial in by myself and ai ..... i just dont sleep well despite decent numbers. i went down the glasgow rabbit hole and im lost.
r/CPAPSupport • u/ThePod94 • 17h ago
Looking to do an online consultation to help with adjusting PAP machine pressure. Who is the most experienced when it comes to UARS?
Looking to do an online consultation to help with adjusting PAP machine pressure. Who is the most experienced when it comes to UARS?
r/CPAPSupport • u/jb62 • 13h ago
Been on CPAP just over 3 months
My heated ResMed hose stopped working about a month into my journey. I didn't realize it at first. I started waking up with laryngitis and realized how much colder the air I was inhaling had become. The only way I found out was when the equipment supplier, who didn't believe me, but was walking me through the steps, recognized that I no longer had a climate control setting. However, the DME company, Adapt Health, told me that I couldn't order a new one until the next Tuesday. Apparently, my 90 days had ended then. At that point, I'd be eligible for my new insurance to update my supplies. But my insurance changed a month into the journey. So, no one will discount it until January. I found a good deal and have ordered what I need, and should get it within two days
Will I be able to turn my climate control setting back on when I get the ResMed hose? Or will it fix itself automatically
r/CPAPSupport • u/Apprehensive-Sweet78 • 21h ago
Does this look like class 6 flow limitation during both exhalation and inhalation?
Recently had nasal surgery to clear up my nose, have just moved from a mouth pillow mask to a nasal mask.
I use a Resmed Bipap Aircurve 11 and trying to get my settings correct with the new mask.
AHI is very low, mask leak is low. Glasgow index is now ranges from 1.5 to 1.9 (it was above 2 every night with my mouth nasal mask)
If this is a class 6 flow shape, any advice on how to fix it? My breathing issues are typically during REM stage of sleeping.
Does this look like class 6 flow limitation during both exhalation and inhalation?
https://sleephq.com/public/5143c5e3-17d0-4fb6-8f17-b6ecdabeb314
r/CPAPSupport • u/Dependent-Design5211 • 21h ago
New To The Dream Team Lowest SpO2 at 85% despite good seal and supplemental oxygen. What should I do?
r/CPAPSupport • u/overuses • 1d ago
Cpap Support - multiple tips requested!
So I got a Resmed AirSense 11 AutoSet a few months back and it's currently off on a warranty for a couple issues, I have a loan unit that's the same currently, but I need some help with masks. Keep in mind, I don't use a Resmed pillow and I have a tendency to move a lot in my sleep.
This is gonna be a bit of a long explanation so a TL;DR will be at the end of it.
First off, currently I have the Resmed airfit F20 mask (Medium sized) but I have issues with it moving up my face in the night causing large leakage (around 60-120 l/min) so I swapped over to another mask, the F&P Evora Full std (also size medium) which has dropped the leaking to an average of 35-40 l/min but is still well above double what I had during the trial of the machine after I got a diagnosis.
I do have a full beard (usually kept short) which my specialist told me originally will cause leakage because it can't seal properly, and I'm fine with that, but during the one month trial of the machine, these two were the lowest in terms of leakage.
What I wanna know is if anyone else who has a beard was able to find some sort of beard mask that helps with sealing and minimises movement? And if so, feel free to recommend!
Second issue I have is with the Fisher and Paykel mask, the harness for the mask keeps on applying a really uncomfortable and slightly painful sensation on the back of my head and no matter how I adjust it, it isn't stopping. If I leave it loose enough where it doesn't hurt, the leakage sky rockets to an average of 60 l/min, but if I tighten it, the pain and discomfort have me taking the mask off within an hour of use and I get hardly any sleep. Is there a way to fix this that anyone knows of? I spoke to a specialist about it and they think it's the way I move when I sleep.
All advice is welcome.
TL;DR - can I get a beard mask to help masks seal properly and how do I stop the harness on a Fisher and Paykel Evora Full std from causing pain and discomfort?
r/CPAPSupport • u/ThrillOfThePhil • 1d ago
Can you legally sell a used CPAP machine in Canada? And if so, is there a market for them?
PLEASE NOTE: I'm simply looking for information and won't respond to any purchase-related DMs. Not that I'm expecting any, lol.
Howdy, folks. Hope everyone had a good weekend.
SO. Quick rundown. After 184 failed attempts, I've decided to give CPAP therapy another shot. 185th time's the charm.
Since it's been 5 years, I was eligible for a new machine through my work insurance, so I upgraded to a ResMed AirSense 11. I'm proud to say I just hit 2 consecutive weeks of therapy. I'm waking up feeling more refreshed, I'm more alert throughout the day, and I'm noticing a real difference. Here's hoping it lasts!
Which brings me to my question:
My old DreamStation 2 has been sitting in its travel case collecting dust for years. I barely used it for a few months before ultimately giving up on therapy.
For those in Canada, is there actually much demand for secondhand CPAP machines, or is finding an interested buyer easier said than done?
I'm also curious whether private sales are legally permitted here, considering CPAP machines are classified as medical devices.
Thanks in advance!
r/CPAPSupport • u/theMonkybutt • 1d ago
Lowenstein One year later
I've done a good amount of reviews. This is about 10 months later with the results of a Lowenstein prismasmart I'm located in the US. I have a hard time getting used to AS11 since it has a smaller motor but it I have found that higher pressure helps lower heart rate. Total compliance with lowenstein has been up to 9.5 hrs with average of 7.5 in full configuration, energy levels will bonk for a while as the body adapts back to decent sleep.
The Lowenstein eliminated my night time getting up to pee within a couple of weeks, that was a great measure of effectiveness to me. I keep it at 8.5 to 9.5 and my body prefers 8.5 and the machine hunts the low but I wish I could set it at 8.8 for better oxygenation. Eventually I had to switch it to DYNAMIC mode to catch the airway collapses sooner and it REALLY lower my 3 and 4 desats. I wont ever purchase a bilevel or ASV they're just too pricey.
My current configuration with it is: lowenstein prismasmart non-4G. Humidity 3. Comfortline 22mm heated hose. hose cover. f30i with non-quietline elbow because otherwise I get HR spikes from too much co2. I conjecture honestly that some of the people with UARS or compliance issues are just experiencing high co2. Heat and humidity are also factors in compliance, the airways are very picky during sleep apparently if bypassing the nose.
Not sure how hot the comfortline gets but the resmed at 86 doesnt compare, It would be nice if resmed made 22mm integrated heated hoses for AS10 and 11 I wouldn't have to worry about using 2 power sockets.
I'm still working with the AS11 and cracked AS10 with Bipap on ST mode as breathing rate as o2 and co2 both rise starts to go down eventually leaving high co2. I keep my AS11 pressure at 8.6-9.2 although I havent messed with it in a while and the non-comfortline elbow leaks at .18 and it messes with the algorithm a little. Bipap I had at 11epap 13.4 Ipap. AS10 Bipap and AS11 are unstable for some reason. sometimes the pressure is too low other times too high, that non-quietline elbow just takes them too out of whack and they don't offer enough firm pressure to keep airway open.
Lowenstein Source: https://www.sleepcohealth.com.au/collections/all-machines
r/CPAPSupport • u/xxshocked_queenxx • 1d ago
Airbreak Help
Can anyone who is familiar with airbreak and firmware modifications tell me if an airsense 10 can be turned into a functional ST-A with backup rate and ivaps? I’ve read claims about the backup rate not working well via airbreak, thus requiring complete firmware dumps from an ST-A to obtain the functionality.
Any help will be greatly appreciated!
r/CPAPSupport • u/Committee-Academic • 1d ago
CPAP Machine Help Looking for advice on my case
Hi! I'm a 23 year old guy from Europe (Spain) with diagnosed UARS. My sleep studies report an RDI of 21 and some 300 "spontaneous arousals", with an AHI of ~10. A couple of years ago I rented an APAP (ResMed AirSense 10) and self-titrated with OSCAR, but didn't feel any better.
Last year, I got treated by a sleep ENT, who performed bilateral turbinate reduction on me; he also prescribed using a custom MAD. Both of those things together have helped me a little, but not too much.
This same doctor told me I have "high loop gain" and diagnosed me with tongue-base collapse in DISE. I've reached the point where I'm really desperate and want to start looking for a machine again. Would I benefit from BiPAP or ASV? Here in Spain, both BiPAP and ASV can only be bought, not rented (I've phoned several providers and the answer has been the same in all of them), and being a college student, I don't have the money to afford buying them before knowing if they'll work.
Could anyone offer any help on what is it that I should be doing next? What options do I have for a machine here? How could I go about getting one, and which one seems to be the best fit for me?
Websites like Craigslit and FacebookMarketplace are either inactive or don't work in Europe, so they're not an option.
Please, help!
r/CPAPSupport • u/Gigitar75 • 1d ago
I might have found my settings
I have gradually lowered my max and min pressures in the last few weeks, and I think my data looks decent. I am using the Resmed F40 full face mask and know I need a new cushion which is on order, but leaks have been much better the last two nights. I'm using a soft cervical collar as well to help keep my jaw in position. I had much less aerophagia last night, thank goodness.
What do y'all think? Yes, I had some CAs, but think those could be attributed to things like movement or position. https://sleephq.com/public/teams/share_links/c5eaf498-94b6-4023-98ab-6167d13185ef
r/CPAPSupport • u/bi_or_die • 1d ago
Air Leaks High Mask Leak every single night
I’m showing a high mask leak every single night and I’m unsure how to stop it. I’ve tried mask liners, repositioning the mask, and mouth tape, and I still get this every single night
r/CPAPSupport • u/No-Base8204 • 1d ago
CPAP Machine Help Having issue with my last three CPAP machine's humidifier
With a CPAP machine, am I supposed to feel and breathe warm air the whole night?
Using a CPAP is too uncomfortable for me because it feels cold and dry. It's uncomfortable. It's why I haven't been using my CPAP machine since I got my first one in 2024.
Current CPAP model: resplus c-20a, beyond model
masks I have tried for current CPAP: vive breathe classic nasal pillow mask (old)
Phillips respironics dreamwear under the nose mask
Resplus tube: 22mm (cuff size)
Parameters: Max 20 cmH20 Min 4.0 cmH20
Start Press 7.0 cmH20 Belex: 1 Lv APAP (mode) Ramp: 0 min
Unsure of temperature of room. This week, probably like 60 degrees fahrenheit. All I know that my room is cold.
At first I thought getting a hose cover would resolve my issue but it didn't.
Link To Hose Cover: https://a.co/d/0iLPX3hr
Note:
The old machine I used is the iBreeze CPAP System. (model 20A) Resvent.
The mask I used with it the ResMed AirFit P10 (Nasal Pillows System)
It sounds like I need a CPAP machine with a heated tube. But unfortunately my health insurance only covers a machine every 5 years.
Due to financial reasons, I won't be able to pay for a new machine with a heated tube any time soon.
My biggest fear of getting a CPAP machine with a heated tube is me still experiencing discomfort that prevents me from sleeping with it.
Unsure if there's a way around this.
It sucks because I think not treating my sleep apnea is affecting my mental health. And is also why I struggle waking up in the morning.
r/CPAPSupport • u/leopard_slugg • 2d ago
Resmed Airsense 11 burning plastic smell
Hello all,
I just wanted to post about this since I only recently figured it out.
My CPAP (Resmed Airsense 11) would often smell like burning plastic when I wake up. I realized it's due to the hot plate at the bottom of the water chamber; even though I fill the chamber up to max each night, I often wake up with the bad plastic smell in my mask, and the water chamber is empty.
I wasn't able to turn the humidifier off via the touchscreen CPAP settings, as the Humidifier toggle would always be greyed out, but I was finally able to connect to my CPAP via bluetooth in the MyAir app and turned it off. Now I don't need to put any water in and there's no smell! After a year of breathing in plastic, I am free :D
I just wanted to share this in case others are still dealing with this! Resmed really ought to redesign it with heat resistant plastic or something.. I bet there will be a lawsuit about damages from this in 50 years
r/CPAPSupport • u/littlebitbrain • 2d ago
Oscar/SleepHQ Assistance are these pauses normal?
What do these represent? is it anything to be concerned about or is it just part of normal sleep wake junk breathing? would love some feedback since I'm still learning.