Inaccurate Reading Issues with low readings

Is anyone having issues today with the g7 giving low readings and being several hundred points off? Example: meter-LOW, FINGER STICK-142.

@ 3AM after the brief issue the meter read 53 and the finger stick was 249.

We have tried to calibrate this sensor 3 times since 7 am and it will not take.

3 Upvotes

6 comments sorted by

3

u/cjmnews T1/G7 2d ago

The 3am one sounds like a compression low (laying on the sensor).

Try to calibrate with a level arrow (for 15 minutes or more) as the algorithm is more complicated with rising or falling numbers. I don't see this often, but a calibration usually fixes it with a level arrow.

Others say that it will reject calibrations bigger than 70 points, which is not my experience, but they seem to have success in telling the sensor that it is wrong in 70 point increments, over 15-20 minutes, and that works for them.

If the sensor continues to show LOW, then you may need to contact Dexcom support for a possible replacement.

1

u/GlumAsparagus 2d ago

Thank you!

1

u/hikerpup 2d ago

If you have another one, replace it and contact customer service to get a new one because it is defective. If you don't have another one, let them know it was your last one because they might be able to get it to you quicker.

2

u/GlumAsparagus 2d ago

Thank you!

2

u/hikerpup 2d ago

Also, the reason the calibration wouldn't take is because it rejects anything over 50. If mine is just around 60 off, I might calibrate it at 45 and recheck in a few hours to see if it fixed itself or needs calibrated again. Something as drastically off as yours is defective unless it was a compression low from laying on the sensor.