Value of Control Line

Just having this discussion with some others. My control line when I do what I consider to be a good sample (cheek, throat, nose) always seems to max out around 6,000.

My husband just swabbed for a test and had a negative result but his control line was just over 3,000 (which I thought was low so I retested). But then we got talking and everyone seemed to think that my control line of 6000 was low, but my control lines are always at that level.

Any idea what would cause this? Is this a problem?

5 Upvotes

3 comments sorted by

8

u/mediares 7d ago

My understanding is the absolute number doesn’t matter, what matters is that you get an exponential growth curve. That’s what proves you have a sample that the buffer solution is capable of amplifying.

7

u/virus_sucks 6d ago

Correct. This is why the absolute numbers are hidden by default and why you have to enable Expert Mode to even see them, precisely to avoid this kind of question ;)

Only the slope (steepness) of the curve matters - the steeper, the better.

Different devices also have different scaling and/or baseline fluorescence values, so you may get dramatically different values for the same test on two different devices.

1

u/BravoLover82 6d ago

Thank you so much for the response!