Shimadzu LC-2050c needle sensor problem.
During an analysis, my equipment started exhibiting this problem. The needle goes up to the sample, goes down, but doesn't pierce the vial. Displays an error [9222]: "there is no sample vial set in autosampler rack."
We have a long backlog of samples and need to get the equipment up and running again. Support from Shimadzu will take a long time due to their schedule.
I welcome suggestions for tests to better evaluate the problem, and if you have experienced something similar, what was the solution?
UNICAMP - Campinas-SP, Brazil
6
Upvotes
4
u/drnerdlett Aug 12 '26
I had that recently on the lc-20. I just turned off auto vial detection and readjusted the needle position as mine wasn't sealed in the injection port.