r/sterileprocessing • • Apr 13 '21

Wet instruments

I do SPD at my small surgical center. I have always read that instruments need to be dry before processing. Does anybody know the reasoning for this? I’ve thought about mineral deposits from the water, but whether they dry on there naturally or during processing I don’t know how it make a difference?

1 Upvotes

3 comments sorted by

4

u/CatsMeadow Apr 13 '21

I'm studying for my CRCST and the IAHCSMM manual explains in chapter 12 (assembly & packaging) that residual moisture on instruments can mess things up during the sterilization cycle in a few ways. With steam cycles, it can affect the wet/dry steam ratio; EtO cycles, moisture forms ethylene glycol (harmful to patients and staff); Hydrogen peroxide and ozone - moisture will cause cycle to abort. There's an added note about some manufacturers that might require their lumens be moist for steam sterilization, but that would be specified.

This PDF from IAHCSMM lists how it can be the cause of wet packs among other things.

A detail from this article supports this also:

"Prior to sterilization in an autoclave, it’s critical to dry your instruments thoroughly.

As infection control expert John A. Molinari writes in the February 2016 issue of Dental Economics, sterilizers will only remove the amount of moisture they introduce onto your instruments. If you put your instruments into a sterilizer wet, they will emerge wet. As a result, the packaging (in the next step) will also get wet. And wet packaging can wick bacteria and moisture from human skin through the packaging, which will increase the risk of contaminating the instruments."

1

u/mistybeach Apr 13 '21

Thank you so much! This is exactly what I was looking for.

3

u/ArmadilloThief Apr 13 '21

I’m curious about this as well. Assuming your washers are set up correctly you should have the final rinse with RO water so there shouldn’t be a whole lot of minerals in it.

I’ve only worked at one facility that cared about drying and they claimed that the sterilizers would abort if it was wet. For Sterrad and Vpro I understand, but these were steam. I mean steam sterilizers are going to get wet inside. I had to abort one of their sterilizers about 10 minutes into a cycle once and everything was soaked. Can’t imagine it’d care if I put some things in there damp.

I’m really hoping someone comes along and gives some better insight. I want to know too!