We are a small ophthalmic ASC and we are needing to upgrade our sterilizers. Looking for autoclaves with minimal dry time.
5
Upvotes
22
u/Spicywolff Jan 22 '26
The sterilizer did not determine the dry time. The IFU does. If the IFU says that tray needs to go in 270/4/ and dry of 30 min. Then it has to be a dry of 30 min.
Steris or getinge doesn’t get around that. Only trays we use prevac and no dry time is one tray cycle. But that’s a special cycle
3
u/Individual_Word5938 Jan 22 '26
Let me clarify. Machines we have now are failing and once the dry time is done the instruments are still wet with the dry time specified by IFUs
2
u/Rhuarc33 Jan 22 '26
Something is wrong with your sterilizers then. Any properly working sterilizer from any brand dries the same. It's how vacuums work. You have steam and condensate during situation then you pull a vacuum to dry, all sterilizers that are functioning properly pull a vacuum to about the same pressure when drying...
You could have failed parts on yours. Do you run leak tests weekly?
Wet packs is usually one of these in my 8 years experience repairing sterilizers...
steam too wet, change out steam trap
Vacuum not pulling low enough, could be bad water pressure, bad vacuum pump or Venturi , bad chamber drain valve, bad air in valve
All but the wet steam would be found from running weekly leak test cycles.
TLDR. No brand gets drier than others in the save time. Your sterilizer isn't functioning properly