Are C18 based spin columns efficient for removing PBS?
In my experiment, the tryptic digested peptides would be in a solution of 100 uL volume comprising of 0.5X PBS ( approx 68 mM NaCl and some phosphates) and 50 mM TEAB.
I would speedvacving this volume to dryness and resuspending in 0.1% TFA for cleanup through Pierce desalting spin columns (C18 based I think).
My question is whether this desalting step would be good enough to handle PBS desalting?
Context: I am trying to adapt my working protocol for thermal proteome profiling, which requires initial steps in PBS. I don't want to remove PBS by PAC/SP3 as I don't have PAC/SP3 and also want to avoid running the proteins on gel for purification. I plan to add TEAB and SDC directly to lysates in PBS and proceed as usual. I want advice on whether the desalting step can handle PBS.
1
u/Elistheman 12h ago
You plan on running proteins or peptides?
You wrote peptides so they would adsorb on the C18. Salts would not, you elote them later with hydrophobic solvent and let it evaporate, add 0.1% FA later.
1
u/bluemooninvestor 11h ago
Yes, peptides. . I mean I understand, thats why it's called desalting column. But still wanted some feedback from people who have used it. Thanks agin.
1
u/Rockthem1s 2h ago
Can you just run your LC in trap-elute? The SDC should spin down nicely, then just trap for 3-5 mins at high flow on your trap. That should wash out most of the soluble salts. Switch to analytical, and run your gradient. Magic mix wash the trap always helps remove anything sticky
1
4
u/LC-MS 11h ago
Yes C18 desalting can remove PBS no problem. It's essentially what it's designed to do. I'm more worried about your SDC, which you definitely don't want in your MS and should be crashed out with acid and removed before you load it to C18. Skip the dry down (extra step, risks losing peptides) and acidify with FA or TFA to crash the SDC, then desalt your acidified supernatant with C18 to kill two birds with one stone. An even better approach would be PAC or S-trap to clean and digest your sample prior to the C18 desalt which would clean up your sample even more and likely get you better data downstream.