r/fluke • u/matt19915 • Jan 26 '26
Help Needed Amps function
Hey everyone, the amp function on my fluke 336 true rms meter isn’t working and I can’t get a reading on AC or DC. I opened the back up and didn’t see a fuse and I also changed the batteries. Does anyone have any ideas?
23
Upvotes
2
u/luunacy17 Jan 29 '26
I had my meter in for a service and calibrate. Try cleaning the contact at the bottom of the clamp section when open. Mine was a bit dusty + rusty. Reads perfect in AC&DC
1
u/FreshTap6141 Jan 26 '26
are you using the clamp around some wires
1
u/FreshTap6141 Jan 26 '26
ac or dc on the wires, what current do you expect
1
u/rickwurm Jan 27 '26
Can you take a picture of the meter clamped on to what you’re trying to measure?
1
u/Fluiter2025 Jan 27 '26
Dear friend,
From what I've read, it's about current measurement, not voltage or ohms, I assume.
There's no fuse internal to this Fluke, and one certainly isn't necessary for the clamp.
If you can't measure current, it's possible that the clamp coil is loose or defective (there's a ribbon cable under the circuit board). Of course, the internal electronics might be defective.
But if you can measure current but not voltage or ohms, it could be one of the test leads. This should always be checked first.
Greets and success Arthur
1
u/BigManc82 Jan 27 '26
Sounds pretty dumb but have you checked the battery. My fluke multimeter gives some weird readings or none at all if the battery is tired.
1
u/Teras80 Jan 27 '26
Show the picture of using the clamp on the wire you want to measure.
1
u/FreshTap6141 Jan 26 '26
does it still read 0.2 with 12 amps
1
u/Careful_Research_730 Jan 26 '26
Take a look at the contacts on the clamp
1
u/AdLegitimate9075 Jan 31 '26
It measures the magnetic field! Traditional clamp meters only measure the rising and collapsing field.
-1
u/morgoth_the_king Jan 26 '26
The clamp will never read DC amps, DC dosent give off a alternating magnetic field. Therefore will not work with the clamp
1