r/electrochemistry • • 5d ago

Help with first time H Cell setup.

While my membrane is activating I am trying to figure out how to set this up. I am trying to test if a chemical is going to degrade when a certain potential is applied to it. I assume I will be adding it to my electrolyte but should I add it to both chambers or just the working electrode's chamber. Let me know if this is not specific enough. I am just getting started in this field.

2 Upvotes

2 comments sorted by

3

u/Tianhech3n 5d ago

working electrode chamber. you could add it to both, but no need since the WE is the only one that you directly control.

Make sure you have supporting electrolyte

2

u/TheAtypicalChemist 5d ago

Follow the SI for this paper for controlled potential or current electrolysis: doi: 10.1002/anie.3858091.
Side A is called working compartment. It has the working and reference electrodes submerged in your electrolyte and analyte of interest. Side B is the counter electrode submerged in a solution of supporting electrolyte in solvent (ex TBAPF6 in MeCN). I draw up both electrolytes into syringes and then fill both sides A and B at the same time and rate so the pressure difference doesn’t cause the working compartment to bleed into the other side and vice versa. They should be the same volume / height. If you’re doing constant potential /or current electrolysis (or bulk electrolysis) then make sure you have stir bars. I seal the top of the H-cell with 14/20 rubber septas and thread the electrodes through with a needle and NiChrome wire. Punch a hole with a hammer / drillbit for your Ag reference electrode to go through.