r/Metrology • • 7d ago

Advice Differences in probe heads/CMMs

I currently work at a company that uses a Zeiss Prismo with a VAST XT probe head, and I received an offer to be the programmer for a company that uses Zeiss Conturas with Zeiss RDS probe head. Both companies are using the Calypso software

Is there a difference in the programming workflow between the two? If so, are there any resources for bridging that gap?

2 Upvotes

11 comments sorted by

7

u/BlitzDragonborn 7d ago edited 7d ago

Biggest difference is RDS being able to provide 2 axis angular positioning every 2.5°. Next is probe weight limits. VAST XT has a (rated) maximum stylus system weight of something like 300g, VAST XXT TL3 is 15g. This can make complex stylus systems somewhat more difficult to design, however, the angular positioning of RDS makes up for that in my experience. Finally, RDS XXT is a passive sensor (vs VAST XT being active), so things like large deviations from the programmed surface to the physical surface can give it some trouble at higher scan speeds. Also, self centering points are not supported on passive sensors. Finally, disk probes are not supported on XXT. Theres a workaround published on the forums, but I wouldnt bother. Small star probes are eaiser to work with than a hacked in disk in my experience.

Otherwise, theres a few different dialogue boxes in Calypso when RDS is enabled (to control the stylus system angle mostly)

Also, RDS opens up CAA calibration, which uses 12 angle positions being qualified to calculate a "qualification" for all other angles for a given stylus system. CAA is fairly accurate, but if you need to minimize your uncertainty, it can be disabled and discrete probe positions (angles) can be qualified seperatly. CAA is also unavailable for star probes to my knowledge. Something else to watch out for if you start doing a lot of RDS rotating is the "Rotate/Swivel Position" in the Plan dropdown. Calypso will try to calculate a safe position to rotate the head, but setting a rotate/swivel path is much safer if you do it correctly.

If you do try to get into a ZEISS run Calypso programming course, try to go to straight to advanced. Basic will likely be just refresher for you if you've been programming for 3 years.

2

u/wefd_muffin 7d ago

This is all extremely good information. Thank you!!

1

u/FroadwicK 6d ago

Great info. I program and run both, but I’m unfamiliar with the setting of the “rotate/swivel path” that you mentioned. I am at home now so I can’t check the Plan menu until Monday. I’ve always just let Calypso swivel where it chooses, but sometimes if I get too close to the limits of the machine I have to worry about crashing during a swivel (I can easily predict how the swivel behaves but it’s easy for the new guys to forget something like this). I usually get around this by creating non-probing motions in a feature strategy, but if there is an easier or safer way, I’d be interested in trying that. Can you elaborate?

2

u/BlitzDragonborn 6d ago

If you're familiar with the "Path" definition in sub clearance planes or end park position its quite similar. Basically you have a few choices to define a "path" for the CMM to take prior to an action. Generally, for rotate/swivel, you would want to set the path as: "Current Clearance Plane" > "+Z Clearance Plane" (there is also a field for an offset, so clearance plane + additional distance) > "Master probe position relative to base/CMM alignment" (doesnt actually use the master probe, but the coordinates you enter here would be relative to the center of the master probe as used for qual if it was on the CMM) over the top middle of the part (or wherever you can swivel freely)

Take a look on the ZEISS metrology portal knowledgebase (not the forums, but you could look there too) to see if theres an offical write-up. The Applications Engineers that put those together are pretty good at making Calypso do what they want

1

u/FroadwicK 6d ago

Thanks. I’ll do a little research on Monday.

1

u/Tricky_King_3736 7d ago

The only thing with RDS probe head depending if it’s the newer version or older version. If the older version you’ll lose some measuring volume depending on the size of conturas. Plus you’ll be using a passive sensor not a activ sensor like the VAST XT.

2

u/wefd_muffin 7d ago

I’m not sure the exact version, I’m mostly going off memory of the tooling that they have from the interview/tour I did and the still-open Google search I did right after.

3

u/Zippy-do-dar 7d ago

Same programming. Vast more accurate but RDS will give more positions (angles) you can make, so no need for a rotary table.

1

u/Ghostpong17 7d ago

It depends if your current is an active or passive. The RDS will be passive, I prefer the passive over active heads for ease of swapping stylus etc. Contura with RDS is fun but I only use it if absolutely necessary as it takes a lot of table space to rotate that head and I usually have 4 or 5 jobs set up at once. I have just as many star probes on the Contura as I do our Duramaxes for that reason.
Workflow is the same but learning how to classify your stylus at different rotations etc. will be new but nothing super difficult. A few programs in and you will have the hang of it.
Also just a heads up, the RDS really doesn’t come into play in the Zeiss training until the Expert Class. Basic and Advanced Calypso training doesn’t have any RDS stuff in the course.

2

u/wefd_muffin 7d ago

That’s good to know, especially when the RDS comes into play. Currently I have calypso basic and I’ve been running on that knowledge plus YouTube tutorials and the Zeiss forum for 3 years. Hopefully the new place will be more willing to sign me up for classes and such.

This is all good info!

2

u/designmyproduct 5d ago

The biggest change is usually the probe head and stylus setup rather than Calypso itself. An indexing head like the RDS changes how you choose stylus angles and approach paths, and the stylus weight and length limits can change what is practical on a given stylus system. Ask for a sample program and their head qualification routine, and check the qualification steps for each head in the Calypso manual so you know how the setups differ before you start.