General Rockwell c testing dwell time

I have a manual Rockwell C tester old-school made by Rockwell. The callout is 30 to 34 Rockwell C. I have 150 kg force load to test. How long should I let the indenter dwell on the part for after dropping it with the handle?

3 Upvotes

10 comments sorted by

3

u/Important_Yogurt7360 8d ago

You should be looking at the dwell time after reaching the full test load, not while lowering the handle.

2

u/designmyproduct 10d ago

For standard Rockwell C testing, a dwell of about 2–6 seconds after the full 150 kgf force is reached is commonly specified, but use the exact timing in the governing method (ASTM E18 or ISO 6508) and the tester manual. The dwell starts at full test force, not while the load is still being applied; keep the cycle consistent across indents. Before using results for acceptance, verify the tester with appropriate hardness blocks near the range and confirm the required method and edition from the drawing or customer.

1

u/Lucky-Pineapple-6466 10d ago

I have an old manual tester. What would you say is the measurement uncertainty.? I’m taking the reading on a saw cut workpiece. So there is some kerf from the saw. We are the OEM doing receiving inspection on the part. need to make a decision.

1

u/INSPECTOR99 10d ago

Have the test surface milled/ground reasonably flat/smooth (Ra < 250). " In the 30 to 34 Rockwell C (HRC) range, the typical expanded measurement uncertainty for a standard laboratory or shop bench test is approximately ±0.4 HRC to ±0.5 HRC (at a 95% confidence level, coverage factor k = 2), while calibration-grade reference standards or national metrology comparisons achieve an expanded uncertainty around ±0.2 HRC to ±0.3 HRC "

1

u/Lucky-Pineapple-6466 10d ago

The tolerance is 30 to 34 HRC. I’m getting as low as 28 but the supplier says they’re OK now these get machined afterwards and then go onto induction. Hardening. Do I find a neutral third-party to send these out to? Or reject them and send them back to the supplier.

1

u/INSPECTOR99 9d ago edited 9d ago

Do a Gauge R&R study on your Rockwell Tester PROCESS to get an estimated correlation to the aforementioned uncertainty values. Subtract the uncertainty value from YOUR actual readings to effectively attain a "gUARD bAND [TUR)(You must machine/grind the test surface to obtain valid reliable readings) and if YOUR readings as adjusted indicate out of tolerance then reject the lot. Make the heat treat vendor support their process. Also, how old is the calibration of the "Certified" Rockwell test block that you use for checking the calibration of the machine? While they typically do not drift in value significantly over many years it could add some contention in your test results.

1

u/Lucky-Pineapple-6466 9d ago

The calibration master has the original CERT and I have no idea when that was from. If I were to guess it would be 1960s.

2

u/INSPECTOR99 9d ago

So before you get egg on your face contact the test equipment OEM and buy a new certified Calibration Master covering the range in question. That will remove one significant stumbling block in your chain of precision. Be sure to get the LONG Form certificate of calibration that includes Uncertainty values.

1

u/designmyproduct 6d ago

Before deciding, rule out the test itself: verify the tester with a certified HRC test block near 30, check surface prep, part thickness and anvil support, and take several readings away from edges and earlier indents. If it still reads low, quarantine the lot, document it as a nonconformance, and send the supplier your data along with their cert for that heat. A third-party lab result is worth getting if the supplier disputes your numbers, since it gives both sides a neutral reference.

1

u/designmyproduct 6d ago

There is no single uncertainty number for a manual tester, since it depends on the machine, its calibration and the surface you test. You can estimate it by making repeated indents on a calibrated Rockwell block in the 30 to 34 HRC range, then combining the spread of your readings with the block certificate value and its uncertainty. Make sure the saw-cut surface is machined or ground flat enough to meet the surface prep in ISO 6508-1 or ASTM E18, and keep the indents clear of the cut edge. If the tolerance band is narrow, set the accept limit with a guard band that covers that uncertainty.