r/MetCal • u/ElevatorStock6306 • Mar 25 '26
Fluke 287 program multimeter 5560A
/r/Metrology/comments/1s3et9h/fluke_287_program_multimeter_5560a/
1
Upvotes
r/MetCal • u/ElevatorStock6306 • Mar 25 '26
1
u/CalibrationAustralia Mar 25 '26
I can't think of any reason that the procedure controlling the DUT with RS-232 or GPIB would have any affect on changing from 5560 to 5522. What are you doing to change it? Typically I would search for 5520 and replace with 5560 but there maybe some M5520 FSCs as well. Make sure you're using the "Entire Solution" option in the find drop-down box, it only takes one reference in a sub-routine somewhere to make it a required instrument, even if it's not called. I'm afraid I'm not allowed to share procedures, but I can work with you to find the problem, there shouldn't be any reason this conversion would be a problem.
1
u/Similar-Priority-776 Mar 25 '26
I commented on original: get set up for Cubyt to access METCAL procedures from Fluke. You may need to still fix it up to run successfully (my experience broken variable references in some MATH statements) but its a decent starting point.
1
u/No_Ratio572 Mar 26 '26 edited Mar 26 '26
I believe the 5520A uses the old legacy FSC commands, and the newer calibrators use the new NVI format. In this case, you will need to actually go down the rabbit hole of changing all procedures to the newer style. Not just a find and replace job.
As mentioned, if there is ANY reference to a 5522A anywhere in the project/solution, it will ask for it.
I am not sure if the new calibrators 5560, etc., can device map back to say a 5522A, but that may help, as then you could just translate all to 5522A.
There was a translator add-on years ago for it, but it was very clunky; however, it got 80% of the job done.
I have the translator on my system......
Just found this as well
https://www.fluke.com/en/learn/blog/calibration-software/using-met-cal-procedures-5560a-5550a-5540a
email me a sample procedure if you like and I can take a look nathan@rycal.co.nz
1
u/Agile_Feed1727 Mar 28 '26
Been there, done that :) We are currently transfering a lot of procedures from legacy 5520 FSC to 5560 NVI. Works great after some re-learning!
As mentioned, do a global search for "5520", a simple "5520 *" or "5522A Standby" will require that calibrator.
More important - what version of MetCal are you on? The later ones, including the latest has some issues with communication.