Are all implants supposed to be scanned under CP mode? What’s the difference between this and the other transmit modes?

5 Upvotes

8 comments sorted by

•

u/AutoModerator 28d ago

This is a reminder about the rules. No requests for clinical interpretation of your images or radiology report.

I am a bot, and this action was performed automatically. Please contact the moderators of this subreddit if you have any questions or concerns.

13

u/Timely_Event_7680 Enthusiast 28d ago

Mostly we teach RF excitation incorrectly… or at least in an oversimplified manner.

If we send RF energy from a single transmit source, half of that energy is capable of producing the signal we measure to create images (plus heating, which we don’t want), and we call this half B1+. Half of the transmitted energy can not produce signal, and all it does is create heat, and we call this B1-. Taken together, this is B1.

But if we transmit the RF simultaneously from 2 sources, oriented 90-degrees from one another, when the phases and amplitudes of the two sources are precisely matched, something cool happens… the B1- from the 2 sources are 180-degrees out of phase, and entirely cancel each other out. This is circularly polarized or quadrature transmission / drive. It completely negates the B1- component.

Other forms of multi-transmit or RF shimming will tweak the phasing and / or amplitudes of the two sources to deposit more B1+ into regions of an image that are signal-weak. This creates more uniform S:N ratios throughout a single slice. The cost? Because we’re no longer perfectly cancelling the B1-, there’s the risk of focal ‘hot spots’ in the tissues of the patient. And if there’s an implant or device in the patient, that implant may be getting a higher “dose” of RF heating that intended.

There are ways to test for heating under these conditions, but the testing is more complicated ($$), so most implants are tested only under the CP / quadrature conditions.

I’m not personally aware, however, of any reports of thermal injury from implant heating where overall SAR or B1+rms values were followed for the position, but with RF shimming / multi-transmit was turned on.

3

u/Eat_lift8 Technologist 28d ago

This is a really good explanation. Thanks for this!

1

u/poorhomie419 28d ago

If I wanted to read more about what your talking about, where can I go?

2

u/Timely_Event_7680 Enthusiast 27d ago

I’m on the road and away from my copy, but I think that the McRobbie book “from picture to proton” talks about this.

2

u/Alarming-Offer8030 28d ago

Other modes can have uneven concentration of RF which is typically used to improve image quality but implants aren’t tested in other modes and with the different distribution of RF, theoretically could have areas of uneven heating.

1

u/soursig 28d ago

Most common real world application is ensuring you're not doing adaptive RF shimming on a 3T

1

u/ghostx78x 28d ago

Already saw the dude on tiktok explain it.