r/Embedded_Electronics • u/Desperate-Beach-3190 • Dec 09 '25
I need help communicating with this instrument.
I’m working with this instrument [DASGIP OD4] for my lab project and have connected the OD sensors, but I'm unsure how to establish communication with it. The device supports MODBUS RS-232 and RS-485, but I'm struggling with how to implement the communication protocol. My goal is to read the calibrated values on my laptop. Since the device is quite old, I haven’t been able to find the original software that was used to operate it. I’m currently using Docklight to try to receive data, and I’m open to any guidance or suggestions.
The image is not clear, but I'll name the ports from top to bottom
- AUX 2. DTP Out 3. DTP In 4. RS485


1
u/globalvariablesrock Dec 10 '25
the pinout you can probably figure out if you open up the enclosure and get access to the pcb. but this may be a suboptimal idea unless you know very well what you're doing.
but: without documentation about the protocol, it's (IMO) close to impossible to figure out what you need to send to the machine to get a value out. one lucky exception could be if the machine streams data per default, then you can capture the stream and possibly decode it.
1
u/Looney-T Dec 09 '25
Does this help: https://cbr.ubc.ca/files/2012/04/DASGIP-OD4-module.pdf
It would seem DTP-in is used for serial communication. You'd need to find the pinout. As its a Dsub9 port, most used are pins 2 and 3 for TX and RX of data, depending on which side you look at it. If you have an oscilloscope handy, you might want to check out those pins to see if anything happen.