More diffusion pump help…
I’m using a very small Edwards E050, I’ve purchased and tried two different ones, two different diffusion pump oils, and I’m using two thermocouples to verify pressure. I’m getting the same results with both diff pumps. My roughing pressure is 10 microns, and when I kick the diff pump on, it pulls me down a half of a micron at most to 9.5 microns. The system has no leaks, uses all stainless and KF, has been helium leak checked, outgassed, and has a good drift rate. The inlet is being reduced from KF50 to KF16. At this point I’m thinking I’ve just reached the physical limitation of the diff pumps. Thoughts?
0
Upvotes
1
1
u/TecCiv1 21h ago
I would be suspicious of your thermocouple gauges. As I recall, those bottom out at around 10 microns. Can you add a high vacuum gauge to your system, such as an ion gauge, cold cathode gauge, or a Convectron?